<div dir="ltr">Hello everyone,<div>Lukas detected a data acquisition failure on our Optima 1 where it skipped one or more scans in the middle of the run. You can see this event in the time state, which records among other things the interval between successive scans. You can review the timestate from any experiment when you open the run in the Import program, or when you download it in the "View Raw Optima Data" data module. The module looks like this:</div><div><img src="cid:ii_mcp5qk470" alt="image.png" width="881" height="625" style="margin-right: 0px;"><br></div><div><br></div><div>In the upper panel you can select to show the time interval, and in this case you can see that there is a longer time interval between scans at around 300 minutes. This means a scan was skipped. It happened on Optima 1. I want everyone running experiments on Optima 1 to please check that for each experiment you run. If you see this problem repeatedly, we need to call service. A small variation in the time interval between scans is normal, but the length of time is significantly increased for this run at this time, and that is not normal.</div><div><br></div><div>Thanks! -Borries</div></div>